Pretty much what the title says - running 8270 TCLP by SPE (Horizon One Pass, with the in-line carbon cartridge) and pyridine recovery in the blank spikes is abysmal: 4-8% on average, enough to frequently fail even our very forgiving control charts. Interestingly it’s usually a bit better in the MS/MSD, more like 15%. The method suggests that the concentration step may be to blame - we use a Horizon Dryvap system and concentrate the three fractions (acid-base-carbon) in the same reservoir at the lowest heat setting.

Anyone else running this method have similar issues or guidance?
andrew ryder wrote:
Pretty much what the title says - running 8270 TCLP by SPE (Horizon One Pass, with the in-line carbon cartridge) and pyridine recovery in the blank spikes is abysmal: 4-8% on average, enough to frequently fail even our very forgiving control charts. Interestingly it’s usually a bit better in the MS/MSD, more like 15%. The method suggests that the concentration step may be to blame - we use a Horizon Dryvap system and concentrate the three fractions (acid-base-carbon) in the same reservoir at the lowest heat setting.

Anyone else running this method have similar issues or guidance?


Have you tried raising the temperature a little and lowering the gas flow during evaporation? I know that works for PAH compounds sometimes. It is counter intuitive, but who knows.

To see if it is cartridge related or evaporation related, extract a blank then spike it just before evaporation and see how it recovers. If you get near 100% then it is the extraction step, if it is low then it is evaporation.
The past is there to guide us into the future, not to dwell in.
Thanks James - I will just have to play around a bit. Hopefully not too much as those SPE disks and cartridges are pricey.

Makes sense for the PAHs, but pyridine is relatively volatile, so I'm worried a hotter coil will just burn it off even further.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry