Hi all. I am running an Agilent 7200A model GC-QTOF with a DB-23 column installed. Getting bad tune results, and when I run samples or blanks, there is a massive column bleed signal at m/z 216,313,366 (the worst but not only). The DB-23 has been conditioned. Its operating range is 40-250, and my program is ramped 40-230 C. The QTOF had preventative maintenance, including source cleaning, last week. The tune could be an unrelated issue.

Any thoughts? Can post screenshots. Let me know what you want to see. Thanks!
Is the end of the column too far into the source? That can cause bleed and tune problems. Also if the AUX temperature that controls the interface temperature is too high it can cause bleed from the last bit of column before the source.
The past is there to guide us into the future, not to dwell in.
Thanks for your reply! The interface temp is 245, which should be ok. It might be that the column is installed too far in. I got a good tune on Friday.

I broadened the scan range and see not only 366 but 419, 493, 546, 599. Mostly steps of 53 mz.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry