Posted: Thu May 16, 2019 4:03 pm
Greetings, I have an Agilent 5975C MSD (G3174A) which had been idle for over a year in another laboratory, so I do not know its entire working history. It holds vacuum and has appropriate air/water values, I cleaned the EI source so the repeller curve no longer looks like a step pyramid and it autotunes tunes well, the filaments look OK.
While looking through all the possible maintenance steps which the unit may need, I have read suggestions in this forum's archives that the RF coils should rarely need adjusting, although doable after watching a Agilent specialist. The manual says that, with both screws being turned equally, the voltage value should be at a minimum value, with typical readings of 70-100 mV for 100 amu and 500-650 mV for 650 amu.
My values are 44 and 297 mV respectively. Is the goal of the adjustment to 'be at the lowest value' or is it to 'fall between typical printed values?' Should I leave well enough alone, or make adjustments?
Thank you
While looking through all the possible maintenance steps which the unit may need, I have read suggestions in this forum's archives that the RF coils should rarely need adjusting, although doable after watching a Agilent specialist. The manual says that, with both screws being turned equally, the voltage value should be at a minimum value, with typical readings of 70-100 mV for 100 amu and 500-650 mV for 650 amu.