Hello! Our lab has GCMS system DANI with MS TOF detector. Currently we have got a problem with it: autotune procedure fails. Intensity of mass 219 is 50. I have verified that PFTBA in ampoule is quite enough. So I do not think it is because of missing PFTBA. But I have noticed that in menu "TOF Status" there were items coloured in red: "Trap Fault", "LA Focus2" and sometimes "Blanker Output". The most interesting that mentioned items become red when I start autotune, but before it they are green. I would address directly to DANI, but they said that any informational support is only for money. So, I will appreciate any support from you.
Thanks in advance.
P.S. I have attached last good autotune report Image, last bad autotune report Image and snapshot of errors observed for TOF Image.