Per the log book that was stored with the instrument, it was last used in late March of 2010, so had been out of service right at years. The last recorded tune value was 1647V on the EM. I fought a high water back-ground on the instrument for about a week, but it had mostly gone away by the time I left work on Friday afternoon(I understand this isn't terribly uncommon on instruments that have been open to the atmosphere for several years, especially since the 5971 doesn't have have separate source and quad heaters to allow for bake-out like the 5973 and later).
In any case, even though the background has settled down quite a bit, the tune seems to still be all over the place. I might get it to tune one morning with good peak widths, good 502 abundance, and have it go to ~1700V. That afternoon a manual tune might show 502 at ~3amu wide, and hitting autotune might have it fail by maxing out the AMU gain narrowing the peaks or EMV getting the 502 abundance up.
I did FINALLY hit "reset defaults" on Friday morning, and even with the EM at 1500V it actually looked quite good across the board. I had good abundances across the board, only had to set the AMU gain to the low 500s to get 502 down to ~.55 or so, and the AMU offset only needed a slight tweak to have them all in that range. Quicktune only made slight adjustments to the manual tune parameters I'd set, and I think ended up with the EMV at ~1550 to get the abundances it wanted(and I think used AMU offset rather than gain to get peak widths to ~.52). When I checked it before leaving, 502 had widened out to ~.65 or so, but everything else looked good and again I brought it back in line by adjusting the AMU gain again.
Still, though, it concerns me that it drifts that much after a few hours. I'll see what it looks like when I get to work tomorrow(I won't be in until late in the day), but it's still drifting more than I'd like if Friday was any indication. Granted I've been spoiled by a 5975 over the past few years-the only 5971s I work with are ones I service independent of my job, and when I was using one regularly back in my undergrad days I didn't even understand the concept of tuning. This is the first time I've been in the position of doing BOTH analysis and maintenance on a 5971(I've only ever done one or the other). In any case, if I pull up the manual tune on a 5971 even a month after a full autotune my peak widths will still be well under .6.
Also, as general information-I didn't get a rough pump with this instrument, but have a massive Varian unit I scavenged from a Saturn 2000 that needs to be sent to surplus. I've used this particular pump for the past year or better with the 5975, and it gave better foreline pressures than I'm getting even with my newly rebuilt Edwards pump that shipped with the 5975. I swapped them so that I could put the Varian on the 5971, and I've seen foreline pressures below 20mTorr at .5mL/min He and ~30mTorr at 1mL/min(this 5890 does have an EPC-it's a bit weird to me to not have a gauge on the front a 5890). I'm waiting on my Ebay-special vac gauge controller to arrive, as that got lost in storage.
Any thoughts on if the tune drifting is something I need to be concerned about, or if it just needs more time to settle down after storage?
One last thing-when I asked about cleaning the quads in the 5975, my predecessor in my job said "Brenda(local Agilent FSE) would probably kill me if I even touched them." From what I understand, the 5973 and have gold-plated rods fused to the ceramic that are basically not ever to be disturbed. The 5971 manual DISCOURAGES cleaning the quads, and says that they require no periodic maintenance and not to disturb their orientation, but that they can be rinsed with acetone, hexane, or freon. I've also seen reference on this forum to the front part of the 5971 quads being susceptible to "burning" much as what happens in the source. Knowing that I could easily turn a working instrument into a pile of junk, is it worth at least flushing the quads with a bit of acetone or should I leave well enough alone?