Can 0.012 N HCL be used in LC-MS? I know HCL shouldn't be used for LC column. This is a old method which is used to analyze Malic acid in our lab. The are two peaks in my pure standard chromatogram. The method is using UV detector at 210 nm, mobile phase is 0.012 N HCL and water as solvent. I wanted to use the method on LC-MS to identify the 2nd peak. Is it safe to use 0.012 N HCL on LC-MS? Will it damage the analyzer? I use a mixture of low concentration of Nitric acid and HCL on ICP-MS to analyze heavy metal in our products regularly. But I don't know that I can 0.012 N HCL on single-quad(Agilent 6100) or not. Thanks in advance
-
- Posts: 139
- Joined: Mon Feb 22, 2010 11:15 am
HCl is a strong acid and is likely to be corrosive for the MS source. I would not use it without asking the manufacturer. In fact, I think I would not use it at all because there are usually alternatives.
An alternative for you would be to manually collect the peak from your HPLC-UV analysis, evaporate it (provided the compound is stable and non-volatile; you can determine this by first reinjecting in your normal HPLC-UV set up), and finally redissolve it in an appropriate solvent for MS infusion.
An alternative for you would be to manually collect the peak from your HPLC-UV analysis, evaporate it (provided the compound is stable and non-volatile; you can determine this by first reinjecting in your normal HPLC-UV set up), and finally redissolve it in an appropriate solvent for MS infusion.
-
- Posts: 517
- Joined: Thu Jul 14, 2016 12:08 pm
See this topic.
If the sole purpose of HCl is reaching an acidic pH of the mobile phase, which it probably is, I suggest to use an additive which is proven to be LCMS friendly.
0.012M HCl has a pH of 1.92. You can approach this with formic acid, or low concentrations of TFA. Using the formula for weak acids (or an online calculator), you can calculate the concentration you need for the same pH. Keep in mind TFA has its own issues (I don't use it), but it's probably more instrumentally friendly than HCl.
If the sole purpose of HCl is reaching an acidic pH of the mobile phase, which it probably is, I suggest to use an additive which is proven to be LCMS friendly.
0.012M HCl has a pH of 1.92. You can approach this with formic acid, or low concentrations of TFA. Using the formula for weak acids (or an online calculator), you can calculate the concentration you need for the same pH. Keep in mind TFA has its own issues (I don't use it), but it's probably more instrumentally friendly than HCl.
3 posts
Page 1 of 1
Who is online
In total there are 55 users online :: 2 registered, 0 hidden and 53 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am
Users browsing this forum: Google [Bot], MSN [Bot] and 53 guests
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am
Users browsing this forum: Google [Bot], MSN [Bot] and 53 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science, @Food_Science_, @Pharma_Science_
- Follow us on Linkedin: https://www.linkedin.com/showcase/separation-science-/