In the midst of developing a method for low-level PAHs by 8270 and I noticed that Restek recommends their drilled uniliner (hole near top, with wool.) What is the purpose of the hole in these liners?
-
- Posts: 52
- Joined: Tue Mar 20, 2018 2:28 pm
Hi,
The uni-liner will minimise the number of active sites in the sample pathway so less adsorption occurs. This allows a better recovery for the active analytes in your sample.
Hope this helps.
The uni-liner will minimise the number of active sites in the sample pathway so less adsorption occurs. This allows a better recovery for the active analytes in your sample.
Hope this helps.
-
- Posts: 326
- Joined: Wed Oct 02, 2013 12:20 pm
- Location: UK
Hello
Please see presentation below:
https://chromspec.com/pdf/e/rk105.pdf
There is good explanation for this liners design.
Regards
Tomasz Kubowicz
Please see presentation below:
https://chromspec.com/pdf/e/rk105.pdf
There is good explanation for this liners design.
Regards
Tomasz Kubowicz
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
The Uniliner attached to the head of the column so it acts similar to a direct injection into the column. The drilled hole is to allow gas flow and pressure regulation to work properly since this is installed in a standard split/splitless inlet. It also allows for the liner to be swept with gas once the end of the splitless phase has occured.
Without the hole the GC can not control the column flow since is it set up to control the pressure after the injection port instead of before the injection port. These are back pressure regulated instead of head pressure regulated like the older direct inlets were.
Without the hole the GC can not control the column flow since is it set up to control the pressure after the injection port instead of before the injection port. These are back pressure regulated instead of head pressure regulated like the older direct inlets were.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 26
- Joined: Wed Aug 29, 2018 10:43 pm
Ahh, I think I get it now - since the capillary column itself seals tightly with the uniliner, you need to create a new opening to the liner to allow split gas flow. Chalk it up to my poor understanding of the mechanics of split injections. Thanks all for the replies.
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science