LALman wrote:
Silver Chloride is slightly soluble in nitric acid. Each time you dilute you get a slight amount more (taking DF into account).
For 20+ years I've run ICP-AES and ICP-MS using a 2%HNO3-1%HCl final dilution matrix. I do my digestions with those acids in that ratio. This will keep at least up to 1 ppm Ag soluble at the instrument in standards and samples. That way chloride interference is not a problem. You also get better recoveries of As, Sb, and Tl in complex matrix with chloride present.