Hi!
Recently we bought used Agilent 7000A triple-quad (7890A GC equipped with G7000B triple-quad detector).
Several monthes ago we faced the problem with triple-quad detector that suddenly goes offline. MassHunter reported that connection was lost, ping from command line could not get a reply from detector (there is no problem with GC pinging). LCD screen on the face of the detector is operating but keyboard does not (detector have no control LED on the front panel).
GC, detector and PC connected via LAN-hub. Any manipulations with LAN-cables as well as resetting the PC did not solve the problem.

After reset the instrument was operational for weeks but the problem has emerged again.

We have found that the only way to reset the detector is to turn off the main switch onto front panel for a several seconds (we could not find no hard reset buttons on the back panel).

It is interesting that detector has never went offline during run or standbye but only during transition modes - inbetween batch sequence posititions or during venting.

What can be the source of this malfunction? And how to get rid of this problem? What else can I check? Is that really the only way to reset the detector?
Hi,
maybe problem with your MS sideboard, in that case, might need a Agilent field service Engineer with diagnostic tool to check what happened with your electronic board
------
GiMiTEC | https://gimitec.com/catalog
------
Image
gaumit wrote:
Hi,
maybe problem with your MS sideboard, in that case, might need a Agilent field service Engineer with diagnostic tool to check what happened with your electronic board

Hi, gaumit!
Thank you for a reply!
As a first step now we are trying to exclude any possible network connection issues. After reset QQQ is working fine. It is very hard to solve intermittent problems.
3 posts Page 1 of 1

Who is online

In total there are 9 users online :: 0 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry