Somebody has some news about the new triple quad that bruker is going to launch??
- CE Instruments
http://www.scionhasarrived.com/ Register here
- CE Instruments
Now launched
Also www.globalfoodtesting.com
Also www.globalfoodtesting.com
-
- Posts: 29
- Joined: Sun Oct 31, 2010 3:15 pm
CE Insruments...)))
Scion is nice in terms of external size and maybe price, but both Thermo and Agilent will laugh for a week in front of Bruker statement 'Ultra high Sensitivity'..)
Just a couple of examples from Bruker Application note:
- Application Note # CA 281751: the SRM signals at 0.1 ppb of HCB, Lindane and Trifluralin are taken as an example. It's well known that these compounds are the more sensitive in a pesticide multiresidue, so why not to show critical pesticides as Iprodion, Acephate, Deltamethrin, etc.. Peak Area of these compounds are extremely low if compared to the amount...and you are injecting 2 ul!!!! Again, how is it possible that SN ratio is similar or bigger than peak area?? It's really funny!!!
-Application Note # CA 281752: how can an istrument be defined as 'ultra high sensitive' when 1 pg OFN in SRM mode has a peak area around 2000 (see figure 1 Scan time vs peak area)?!? Consider than a TSQ Quantum XLS has a peak area of 6000 with 0.1 pg OFN (and a SN better than 2500:1)
Tell Meredith to put attention to what is published, because both competition and expert customer will discover it!!
Scion is nice in terms of external size and maybe price, but both Thermo and Agilent will laugh for a week in front of Bruker statement 'Ultra high Sensitivity'..)
Just a couple of examples from Bruker Application note:
- Application Note # CA 281751: the SRM signals at 0.1 ppb of HCB, Lindane and Trifluralin are taken as an example. It's well known that these compounds are the more sensitive in a pesticide multiresidue, so why not to show critical pesticides as Iprodion, Acephate, Deltamethrin, etc.. Peak Area of these compounds are extremely low if compared to the amount...and you are injecting 2 ul!!!! Again, how is it possible that SN ratio is similar or bigger than peak area?? It's really funny!!!
-Application Note # CA 281752: how can an istrument be defined as 'ultra high sensitive' when 1 pg OFN in SRM mode has a peak area around 2000 (see figure 1 Scan time vs peak area)?!? Consider than a TSQ Quantum XLS has a peak area of 6000 with 0.1 pg OFN (and a SN better than 2500:1)
Tell Meredith to put attention to what is published, because both competition and expert customer will discover it!!
-
- Posts: 218
- Joined: Thu Mar 31, 2005 5:33 pm
- Location: USA
hermo71,
Seems you are really from Thermo.
IMHO, the default way Thermo Xcalibur calculates its S/N is the funniest one: even when the peak is buried in the noise, you still get three digit S/N.
Seems you are really from Thermo.
IMHO, the default way Thermo Xcalibur calculates its S/N is the funniest one: even when the peak is buried in the noise, you still get three digit S/N.
-
- Posts: 29
- Joined: Sun Oct 31, 2010 3:15 pm
No, I'm not from Thermo...but I have three TSQ Quantum XLS. I know the performance of this instrument for a lot of compound classes (pesticides, PAH, PCB, etc..) in different matrix (food, soil, plasma, etc..)
I have also an Agilent 7000b working fine and two DFS. I had a Varian triple quad GC (sent immediately back)...
So, I'm enough expert in GCMS to say that Bruker GCMSMS is still far away. By the way Bruker sales rep has already proposed me a demo...we'll see..)
I have also an Agilent 7000b working fine and two DFS. I had a Varian triple quad GC (sent immediately back)...
So, I'm enough expert in GCMS to say that Bruker GCMSMS is still far away. By the way Bruker sales rep has already proposed me a demo...we'll see..)
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
- Location: Alamo, CA / USA
That statement would be *much* easier to believe if his email address were not at "thermofisher.com". That's a violation of Forum policy; we encourage vendor posts, so long as their affiliation is clear. He has been banned.No, I'm not from Thermo...
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 1
- Joined: Tue Sep 06, 2011 7:11 pm
We want to purchase a single quad GC-MS. Is anybody using the Bruker Scion? I would like to know how it is performing. Thanks!
-
- Posts: 3
- Joined: Mon Feb 27, 2012 3:56 pm
hermo71, I am From Chile, I have a TSQQuantumxls and have problems when do the tune. It show me the message "INSUFFIENCIENT SIGNAL, OPENING RESOLUTION FURTHER". The message are for both Q1 and Q3.
I cleaned the source and loaded new PFTBA. However the problems persist. Then I carry out a Diagnostic test, all component passed it.
I called to technical support in my country. They suggest that source is dirty.
I will be very gratefully if you send me any suggestion for this problem
Thanks
I cleaned the source and loaded new PFTBA. However the problems persist. Then I carry out a Diagnostic test, all component passed it.
I called to technical support in my country. They suggest that source is dirty.
I will be very gratefully if you send me any suggestion for this problem
Thanks
-
- Posts: 2
- Joined: Mon Apr 06, 2020 10:19 pm
We bought Scion GCMS for Toxicology analysis, it’s good and sensitive, software is user friendly.
Contacted representative for expansion and spare parts, they ignored us, after many calls and emails we got double price for GCMS and expensive spare parts.
No proper communication with them, customer will be the weakest point and will be missed between Scion or Techcomp or Regional office.
Don’t expect any after sales service at all, don’t contact the management or manufacturer will get nothing.
Only point they will run after you at the beginning then will leave you alone, we are requesting buyback offer from other companies after we fed up with them.
Contacted representative for expansion and spare parts, they ignored us, after many calls and emails we got double price for GCMS and expensive spare parts.
No proper communication with them, customer will be the weakest point and will be missed between Scion or Techcomp or Regional office.
Don’t expect any after sales service at all, don’t contact the management or manufacturer will get nothing.
Only point they will run after you at the beginning then will leave you alone, we are requesting buyback offer from other companies after we fed up with them.
-
- Posts: 2
- Joined: Mon Apr 06, 2020 10:19 pm
jeb wrote:
We want to purchase a single quad GC-MS. Is anybody using the Bruker Scion? I would like to know how it is performing. Thanks!
We bought already very good system, sensitive and stable, but very bad representative from regional office up to factory, will be missed between Scion, Techcomp, Regional office.
Must depend in your self for any software/hardware issues.
11 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science