detector range 5973 vs 5975
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
4 posts
Page 1 of 1
Just a quick question. On my 5973 I am trying to fix I notice the detector maxes out about 5-10E6 counts. On my 5975 the detector maxes out at 3.5E7 counts. Is that normal?
-
- Posts: 693
- Joined: Mon Mar 28, 2005 5:38 pm
They each came in several flavors and with different software so I would expect different clipping thresholds.
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
I have had E^7 counts before on 5973's so could be software/firmware versions or possibly EM volt settings, the higher you have to run the EM volts the less linear dynamic range you will have.
I like to try an keep my normal calibration range at a max peak height in the 1 million range just so when I have high junk in the background I still have good peaks for the targets without far of clipping right when a target elutes.
I like to try an keep my normal calibration range at a max peak height in the 1 million range just so when I have high junk in the background I still have good peaks for the targets without far of clipping right when a target elutes.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 1108
- Joined: Wed Dec 22, 2010 10:17 pm
- Location: Chicago
Yea I'm running it a lot this week and the sensitivity seems pretty similar to my 5975 which has a very old (OEM 12 years) detector that runs at 1900ev. The difference in scale kind of threw me off while I was diagnosing it but reviewing old data that is normal for that instrument.
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science