Compared to using the quick swap using the purge Union in GC MS method is resulting in increased retention time, awful peak shape and significant loss of sensitivity. I can not locate any specific guidance on use of the purge Union and would much appreciate any suggestions on why the chromatography has deteriorated and any pointers to manuals on using the purge Union. Thanks.
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
Which way do you have it set up? Are you using it simply as a No-Vent fitting just before the MS, or for back flushing which has more column after the union?
The past is there to guide us into the future, not to dwell in.
-
- Posts: 1108
- Joined: Wed Dec 22, 2010 10:17 pm
- Location: Chicago
From what I understand the quickswap pretty much is a purged union where the union is at the transferline. The quickswap if fed helium at 4psig from an aux EPC and resistance is proveded by a 17cm .1mm ID capillary through the transferline.
-
- Posts: 5
- Joined: Tue Apr 17, 2018 11:45 am
Thanks for your replies.
The set up is for No-vent fitting. The analytical column is installed directly into the union inlet and there is a 2.4 meter length of column from the union outlet to the MS. Is this the correct set up for No-vent or is there too much column length after the union?
The set up is for No-vent fitting. The analytical column is installed directly into the union inlet and there is a 2.4 meter length of column from the union outlet to the MS. Is this the correct set up for No-vent or is there too much column length after the union?
-
- Posts: 517
- Joined: Thu Jul 14, 2016 12:08 pm
Sounds normal. The length of that piece isn't really critical, it can be a meter off without impacting anything.
In our post-column backflush GC-MS/MS system we use 25m x 0.25mm analytical, and 2.5m x 0.18mm deactivated fused silica to the MS. The narrower the last piece, the shorter it can be while maintaining its function. I wouldn't make the diameter difference before and after the union too large.
In our post-column backflush GC-MS/MS system we use 25m x 0.25mm analytical, and 2.5m x 0.18mm deactivated fused silica to the MS. The narrower the last piece, the shorter it can be while maintaining its function. I wouldn't make the diameter difference before and after the union too large.
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
If the transfer line is too small diameter and too long, you will have more restriction in the last part of the flow path and it can interfere with the calculated flow of the column. Also if the union flow control is leaking and pushing flow into the union during a run it can cause the column flow to be low.
I have a setup with two columns installed going into a Tee fitting just before the MS. The first time I tried it I was using a piece of 0.1 transfer line and two 0.25 columns. I set the active column to 1ml/miin flow and the inactive column to 0.3ml/min flow and the peaks look horrible and the retention time was about double what it should have been. I switched the transfer line to a piece of 0.53 column and everything was perfect. The inlet pressure is set assuming hard vacuum at the end of the analytical column, if there is actually pressure there caused by a restriction then instead of 1ml/min you will have maybe 0.5ml/miin( mine was off about that much). This will cause the peak shape and retention time problems.
The restriction could also be caused by a piece of ferrule stuck in the union, if the column happened to shave off a small flake of ferrule material and it came out into the union after install and turning on flow.
I have a setup with two columns installed going into a Tee fitting just before the MS. The first time I tried it I was using a piece of 0.1 transfer line and two 0.25 columns. I set the active column to 1ml/miin flow and the inactive column to 0.3ml/min flow and the peaks look horrible and the retention time was about double what it should have been. I switched the transfer line to a piece of 0.53 column and everything was perfect. The inlet pressure is set assuming hard vacuum at the end of the analytical column, if there is actually pressure there caused by a restriction then instead of 1ml/min you will have maybe 0.5ml/miin( mine was off about that much). This will cause the peak shape and retention time problems.
The restriction could also be caused by a piece of ferrule stuck in the union, if the column happened to shave off a small flake of ferrule material and it came out into the union after install and turning on flow.
The past is there to guide us into the future, not to dwell in.
6 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science