Trouble with Perkin Elmer Clarus 500 GC/MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello Community,

I have trouble with my Clarus 500. Bevor I describe my problem i´ll give you the specs.

i have an Clarus 500 GC/MS with an TurboMatrix 600 ATD injector.
The capillar goes through the back of the gc right into a swafer and from the swafer to the MS-Detector. The PC has an old winXP OS (Home Edition i think).

And now my Problem:

For the beginning i had to do some maintenance at the cooling trap from the ATD (nothing Special) but when i want to restart the system the WinXP was freezed so i had to reset the PC. And here starts the trouble...
After restarting the PC the config.dat from TurboMass 5.4.2 was crashed. I reconfigured the GC in the TurboMass Software. After the reconfiguration I got the Error Message: "-984 0ERROR 3 Value out of Range ERROR AT A: inst. reports error". After countless reconfigurations whit and whitout the help from the Perkin Elmer Customer Service, I reinstalled the TurboMass Softwar.

Furthermore the Clarus 500 GC was report an error too. Error Message from Clarus 500: "Car2 inlet press < 50 psi". The Customer Service told me that the GC want an pressure on the injector (which ist not activated). The Problem was solved while I changed the option for the PPC-Module in the configuration at the Clarus 500. The PPC-Modul wanted a He-Flow at the injector a and b. I changed both to "None" (Note: The injector themself are already on "None"). After restart of the Clarus 500 there are no more Error Messages.

But when I now want to start a sample I get a new Error Message in TurboMass: "-984 SET PUNIT = 23 ERROR 73 AT A = Instrument Hardware is incorrect for requested Setup."

Thats the current status. I hope someone can help me.

Best regards
H. Schmudde
Hello everyone,

the Problem was solved. The customer Service couldn´t tell a solution and liked to come over to substitute the communication board from the Clarus 500 GC. But that wasn´t neccessary. After a reboot of the System (PC & GC-MS) the configuration of the GC in TurboMass was updated and now they work togehter again.
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry