Hi all

I have just moved our GCMS to a new lab and am running some blanks to see if there are any contaminants in the system. I know we need to see a clean baseline with no peaks and only a low level of bleed, but I am interested to know which ions are "typically" seen in the background of a blank sample and in what ratios. For example, in a blank run I typically see m/z 36, 40 and 44 with insignificant amounts of other ions. I'm not scanning below 30 m/z so I might be missing anything lower than that. I'm using helium as carrier.

So, what ions should normally be seen in the background and is this consistent across all systems?
A lot of contaminants are indeed common across different systems.

I can't link the document since it's a pdf, but if you google "gcms common contaminants", first link is an Agilent document (from Researchgate) with a nice table listing m/z and sources.

For example, you'll find argon (40) and CO2 (44) on most systems.
Rndirk wrote:
A lot of contaminants are indeed common across different systems.

I can't link the document since it's a pdf, but if you google "gcms common contaminants", first link is an Agilent document (from Researchgate) with a nice table listing m/z and sources.

For example, you'll find argon (40) and CO2 (44) on most systems.


Yes, I have looked at that document before and it's a useful guide. It doesn't list anything with m/z 36 - any ideas what this could be on my system?
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry