My work around at present is to step the unit into dry purge and block the vent (putting unit in Bake out also works). This works to get the pressure. Then I use reset to abort the run and go back into manual mode and select "C Leak Test". Then I hear a click (vent valve closing?) and I can unblock the vent and see if the unit looses more than a couple pounds of pressure over 5 minutes.
Has anyone had this problem with the "Feed Pressure" function not working? I have some new solonoids and can swap them in if needed. But I don't know which one. The vent valve seems to work and the unit operates in tandem with my Archon 4552 and 5903/6890 GC-MS without apparent problems. My service guy says over the phone that it could also be a corrupted bios making just that menu selection unavailable.
Is there a secret service menu that can be used to trigger the solonoids one at a time for testing?
Any thoughts?