Labsolution - no chromatogram in Chromatogram View

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hello
In Postrun application when i open data file from LCMS analyses I don't see chromatogram but an image of current from MS detector. On the other hand report file looks normal (chromatogram, current image from MS, each peak has MS spectrum etc).

View from my application:
https://prnt.sc/jf6yht

View from help file:
https://prnt.sc/jf6zgl

It is really hard to operate without chromatogram - in help file from Labsolution in each screenshot there is chromatogram.

Thank you in advance for a solution!
It looks to me as though the help-file is showing a total ion chromatogram (TIC) derived from a SIM method, where the TIC is the sum of a handful of ions, and the sample is probably a clean standard containing only those ions, so the TIC is nice and clean.
In your case, the TIC is just showing the baseline noise I'd expect in full-scan data, where you're summing a lot of stuff. You have various options:
(1) right-click on the chromatogram and select "MS data view parameters" (if it's not there, look for "Fragment table", but I think you're on Shimadzu's newer software). Now, if your data were collected in scan mode, you can specify individual masses for extracted ion chromatograms, which will be a lot cleaner. Turn off the TICs (check-box) so they don't dominate the Y-axis scaling.
(2) You can also select base peak chromatogram by entering BPC instead of the mass, if you want.
(3) If you collected data in SIM mode and the bad chromatogram is because one of the ions is present as an abundant background ion, you should see the ions as drop-down options in the MS data view parameters so you can still inspect the other ions and make nice chromatograms for them.
(4) If you collected scan data, then next time you can also set up a SIM event to collect cleaner data, and display a TIC from this event.
(5) If you are reporting on specific peaks and have set up a processing method, then the results section, bottom left, will display the data for the currently-selected compound in the compound table. Choose the compound from the "compound" tab in the method-view, and its data, including a zoomed-in chromatogram, will appear on the "compound" tab of the results view.
Instead Postrun I have used Browser and it works like suppose to do.

Thank you anyway!

Cheers!
Just an extra: in case you haven't already discovered this, Browser also has the advantage that you can overlay PDA and MS data in a single chromatogram, which Postrun cannot; open PDA data in an existing MS chromatogram, and it appears. Of course this also allows you to overlay data from multiple files too. Browser in LabSolutions is much, much better than in Shimadzu's previous software.
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry