I am trying to analyze octane by GCIRMS but instead of one octane peak I am getting several small peaks. I have tried adjusting the inlet temperature ranging from 250 to 150 and my current temperature program is 40C increased to 100C at 4C/min. Does anyone have any successful octane methods?
Just pure octane? Octane in something else? Seems odd. Octane (pure solvent) should be like a model system.

Can you inject any other pure solvent and see a single peak? What are your scan parameters in the MS part of your method? Make sure you're at least scanning up from 50. I mostly do volatiles by SPME in a high-ethanol matrix (MW = 46 g/mole) so I generally go 50-350 when I'm operating in scanning mode. You can go wider or more narrow depending on your application.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry