When I go to 100% ACN (say 5 to 100% ACN in 15 mints and keep 100% ACN for 5 minutes, in a ghradient) in a Trap I always see two predominat ions, 256 and 282 (ESI, +). I normally use SB C8 or C 18 column.
The trap we are using is Agilent SL.
Its is very problematic, specially if the compound of interest elutes out at 100% ACN because these two peaks (282, 256) are so huge that they almost mask other ions.
I have tried cleaning procedure elaborately and it didnt help. The peak is not coming from LC as it is present when I use syringe pump for direct infusion as well.
The agilent engineer is saying it is true universally. Everyone sees those ions when you go go to 100% ACN and keep the gradient for few minutes. Somehow I dont agree with that and I would really really appreciate the readers observation in this regard.
Peak 282 is said to be typical from plasticizer. I guess 256 is associated with that, too, although the difference is 26 which I have no idea about (C2H2???). Can anybody have more ideas?

I have a list of "Common LC/MSD contaminants" from Agilent, dated from 12/2000. According to the list, 256 and 284 come from nylon filters (sometimes used to filter mobile phases). Not sure about 282.

Hi MG,

I also have an MSD Trap from Agilent, and it would be very usefull to have the list of common LC/MSD contaminants.

Did you downloaded from the agilent web page ?.

Or you could send to us by e-mail ?

Thanks

Thanks for all the replies. Any of you see these peaks when you go to 100% ACN and keep it for few mints.
There was a discussion before and some one attributed 282 is the protonated oleic acid amide(281+H+)
The common LCMS contaminant table can be obtained from Agilent
Thanks again.
Hi, MG,

I am very interested to see the list of "Common LC/MSD contaminants" you mentioned. Is that possible for you to post this list? Thank you very much.

Jan

I don't remember where I got the list. It is a photocopy, probably dating from when we had an LC/MSD. I couldn't find it on the web either. I imagine Agilent would probably make it available to their customers who use their LC/MS equipment. I don't really want to post the whole list here, mainly because I don't have time to type it all in, and also because I suppose there could by copyright issues. Sorry.

In the absence of that, here's a link you may find useful, that gives several references on contaminant ions.

http://www.xasa.com/grupos/en/sci/artic ... .mass-spec

I already had the U. of Southampton table and the JASMS article printed. Neither one mention 256, but the JASMS article attributes m/z 282 to "plasticizer in polyethylene".
7 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry