- Greg Cawthray
I am attempting the analysis of low weight organic acids after MTBSTFA derivitization on GCMS (HP 5890 and 5971MSD). Column is a HP5-ms, 0.25mm*30m, 60C 2min, 60-150 @15C/min, 150-300 @6C/min, 300 1min, carrier He at 45ml/min
I am relatively new to this application, and have a few queries
1. do i need to condition the column with (say) silyl-8 to reduce the affect the MTBSTFA has on the reactive sites?
2. injection of 1uL splitless of the MTBSTFA alone gives me many peaks all thru the run, is this possibility interation of the MTBSTFA with the stationary phase?
3. injection of MeCN sees no "ghost"peaks as seen in MTBSTFA, so i ahve assumes injections port contamination is not an issues.
If someone can offer some advice, i would be greatly appreciative!!
Regards
Greg