Does any one knows a method for " Determination of Ethylene Glycol in aqueous environmental sample using GC-MS"
I have found one method which require deriviatazaion. I want t directly inject into GC-MS
I have done it before direct inject but it isn't the best thing for the MS injecting water. I was using an old HP5995 which had solid metal quads.

For best separation you will want to use a Wax column, Restek makes a StabilWax-MS column that has very low bleed which should work.
The past is there to guide us into the future, not to dwell in.
Extracting it is nigh on impossible. We do what James does; direct injection onto a polar phase. Works for mono, di and tri ethylene glycols.
Mark Krause
Laboratory Director
Krause Analytical
Austin, TX USA
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Majestic-12 [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry