Hello! I'm new to the forum and have a question regarding the use of counterions in reverse phase LC-MS. Based on the literature I've read, I believe this could improve my peak shape, but I have no experience in this field. What type of counterions can I use when working with MS detector? Also, which concentrations are normally used? Any help or literature recommendations on this topic would be appreciated.
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
- Location: Alamo, CA / USA
The key requirement for LC-MS is that *everything* in the mobile phase must be at least somewhat volatile. To grossly oversimplify, that means use organic rather than inorganic ions. Concentrations typically will be in the 5 - 10 mm range.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 404
- Joined: Wed Nov 09, 2016 12:43 am
As Tom pointed out you are limited in the buffers you can use because of volatility. 0.1% Formic Acid is a typical acidification agent (~pH 3). You can use Ammonium Formate if the target molecule likes pH 7.
DO NOT use TEA (Polysorbate, Tween...) or other compounds that are likely to contaminate you MS source!
DO NOT use TEA (Polysorbate, Tween...) or other compounds that are likely to contaminate you MS source!
-
- Posts: 88
- Joined: Mon May 16, 2016 11:50 am
most MS setups are MPA: water with 0.1% formic acid and MPB: as ACN with 0.1% formic acid. This is a good start with a C18 column
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science