Hello Forum!
On an Agilent GCMS system (7890A/5975C) we weekly check if the recommended PFTBA- fragmentation conditions (69mz: 100%, 219mz:~50%, 502:~1%) are met, which was the case for months.

After cleaning the ion source and a following tuning procedure however, the PFTBA fragmentation pattern changed and now turns out to be: 69mz:100%, 219:~90%, 502:~10%.

Has anyone experienced this phenomenon before?
environeer wrote:
Hello Forum!
On an Agilent GCMS system (7890A/5975C) we weekly check if the recommended PFTBA- fragmentation conditions (69mz: 100%, 219mz:~50%, 502:~1%) are met, which was the case for months.

After cleaning the ion source and a following tuning procedure however, the PFTBA fragmentation pattern changed and now turns out to be: 69mz:100%, 219:~90%, 502:~10%.

Has anyone experienced this phenomenon before?

After cleaning the source you gained proper sensitivity for high masses.
Also if there is still some water hanging around in the analyzer it will bias the high mass abundance. I usually let mine pump down overnight before tuning, and then after a week do it again just in case. It takes a while for all that metal to come to a thermal equilibrium, but once it does it holds for a long time.
The past is there to guide us into the future, not to dwell in.
Solved!

It turned out that a wrong/different tune file caused these issues. stune and atune were somehow mixed up and falsely attached to the current method.

Air/Water checks are performed weekly, and indeed, after maintenance on the MS, an adequate amount of time is given for proper equilibration.

Thanks.
environeer wrote:
Solved!
It turned out that a wrong/different tune file caused these issues. stune and atune were somehow mixed up and falsely attached to the current method.
...

But you don't invoke tuning procedure from analytical method, do you ?
Then, no matter what tune file is attached to the analytical method you may stune or atune instrument.
dblux_ wrote:
environeer wrote:
Solved!
It turned out that a wrong/different tune file caused these issues. stune and atune were somehow mixed up and falsely attached to the current method.
...

But you don't invoke tuning procedure from analytical method, do you ?
Then, no matter what tune file is attached to the analytical method you may stune or atune instrument.


Could be the tune file that the method uses for operating parameters.
The past is there to guide us into the future, not to dwell in.
dblux_ wrote:
But you don't invoke tuning procedure from analytical method, do you ?
Then, no matter what tune file is attached to the analytical method you may stune or atune instrument.


No. The tuning is done manually.
7 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry