- Posts: 8
- Joined: Wed Apr 26, 2017 3:27 pm
Injection Volume 500 ng
Flow Rate 0.4 mL/min
Solvent A 0.1% formic acid in LC/MS H2O
Solvent B 0.1% formic acid in acetonitrile
Wash & Purge Solution 50% Methanol
Seal Wash 10% isopropanol
Column Acquity UPLC BEH C4 1.7 µm, 2.1 x 100 mm column
Column Temperature 80˚C
Running time 10 min
Time (min) Flow Rate (mL/min) % Mobile Phase A % Mobile Phase B
Initial 0.4 97 3
5 0.4 10 90
6 0.4 10 90
6.1 0.4 97 3
7 0.4 97 3
Drying Gas Temp (˚C) 270
Drying Gas Flow (L/min) 14
Nebulizer (psi) 20
Sheath Gas Temp (˚C) 400
Sheath Gas Flow (L/min) 12
Capillary (V) 5000
Nozzle Voltage (V) 2000
iFunnel parameters H pressure RF: 200
L pressure RF: 100
MS TOF Fragmentor 365 v
Oct1 F Vpp 750 v
It may be the formation of salt and solvent clusters with my protein. How can I minimize or remove this effect? I tried using 0.1% formic acid (mobile A) to load samples instead of water. It helped, and the first peak become higher, but there is still the 2nd peak.
I am quite new for MS, and I have a lot to learn. This is the first big problem I encounter for me and my team. I hope that I could get helpful suggestions and input from experts here. Thanks in advance!
David