Shimadzu QP5000 GCMS Vacuum Parameters
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
5 posts
Page 1 of 1
I have acquired a Shimadzu QP5000 GC/MS from my former employer. I used this instrument 10 years ago and have forgotten most of what I knew. I am trying to get the instrument up and running, but cannot get the turbo pump to start. My question is what level vacuum at the fore-pump is necessary for the turbo pump to start? The Vacuum monitor in the Vacuum Control Window says the "Lower Vacuum" is 4.2e+000 Pa. Is 4.2 Pascal sufficient, or should I be getting lower values than this? BTW I checked power to the turbo pump and there is none there, so I am thinking something is keeping it offline
-
- Posts: 418
- Joined: Thu Sep 25, 2008 3:40 am
jayefield wrote:
I have acquired a Shimadzu QP5000 GC/MS from my former employer. I used this instrument 10 years ago and have forgotten most of what I knew. I am trying to get the instrument up and running, but cannot get the turbo pump to start. My question is what level vacuum at the fore-pump is necessary for the turbo pump to start? The Vacuum monitor in the Vacuum Control Window says the "Lower Vacuum" is 4.2e+000 Pa. Is 4.2 Pascal sufficient, or should I be getting lower values than this? BTW I checked power to the turbo pump and there is none there, so I am thinking something is keeping it offline
In GC/MS, 4.2 Pascal is about normal pressure that the foreline pump can get.
-
- Posts: 41
- Joined: Sat Jan 21, 2012 11:43 pm
Turbo pump controller uses contact closure to start pump. Thus you can override/bypass.
-
- Posts: 3
- Joined: Wed Apr 26, 2017 9:28 pm
Thanks, Thohry. Good to know fore-pump vacuum is ok.
Alexei, where is the contact closure that can operated manually? only manual controls are for the GC. There are none on the MS unit.
Alexei, where is the contact closure that can operated manually? only manual controls are for the GC. There are none on the MS unit.
-
- Posts: 41
- Joined: Sat Jan 21, 2012 11:43 pm
The masspec monitors for-pump vacuum and commands (via contact closure) the turbo controller to start spinning. The connections are at the back of the turbo controller. Don't recall the details, sorry.
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science