The vacuum pump for our Agilent 5973 Mass Spec is leaking oil. Can I safely use an older Edwards 2 stage E2M2 pump till the E2M1.5 pump is rebuilt.
Thank you in advance
Chuck
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
benhutcherson wrote:
I think Agilent likes the quiet little E2M1.5 pumps, but it seems like they all start leaking oil and have other issues eventually. They run incredibly hot, and from what I've seen a get a higher hydrocarbon background in my 5975 than even my 5971(which uses a massively oversized Varian DS102).
The last time my E2M1.5 sprung a leak, I got away with just a replacement of the big gasket between the pump body and motor. In a few years, it will probably be full rebuild or replacement time(our guy at work who normally does that stuff looked at the tiny replacement vanes I had and said "no way-send it off"-we do have a pump shop in town fortunately).
My Varian 300-TQ has an Edwards RV3 on it, which I have zero complaints about. With the turbo(which helps a lot) backed by that pump, I see low 10^-7 torr or sometimes even high 10^-8(single quad EI mode-it hops up to high 10^-6 when I'm doing CI with CID). One independent service engineer I've worked with says he always keeps a new/rebuilt RV3 in his trunk for when he's going to sites. He says he won't replace like-for-like on a failed E2M1.5, the Pfeiffer equivalent, or the new scroll pump-he puts the RV3 on it and calls it a day.
The E2M2 is still a great pump. I'm a big fan of the Varian DS102, which Agilent still sells branded Agilent, also. That was the pump Varian would ship with everything from their budget ion trap models(which generally don't need as hard of a vacuum as quads) and even up to some GC-TQs(the LCs and convertibles on the same basic design used either two pumps or sometimes one really big one). It's a step up from the E2M2, and IIRC without looking now in some specs is better than the RV3 and in some is worse.
As an aside, the 5971/72 are capable of Ci(with a proper source/.gas manifold) and they were only available with diffusion pumps. The 5973 and later require a "performance turbo" for CI use. I posted the question as to why to-again-a former HP/Agilent CE/FSE who is now independent, and he gave me an answer that makes sense even though I probably wouldn't have thought of it. The E2M2 backing a diffusion pump is able to handle the extra flow from CI just fine, while the E2M1.5 is just not up to the task. That's also why Agilent won't pair a diffusion pump with a scroll pump-the pair can't get a good enough vacuum even for EI.
BTW, my 5871 that's now a 5972 reliably pullls high 10^-6 with analytical flows through a .18mm column when backed by the DS102. I have everything I need to do CI on it(source, manifold, plumbing to connect the two together) but even then it seems a bit jerry-rigged and CI is so much easier and more versatile on my Varian 300. Even so, I think it could easily handle doing CI.
James_Ball wrote:
The main reason for the E2M1.5 was the smaller form factor so they would slide up under the MS and save bench space. I would rather deal with the larger pump sitting beside or below the instrument just to get the extra pumping capacity and lifetime.
I have the RV5 on my 7000C QQQ and it started leaking after a couple years. I still have it running just it does its own oil changes every few months. There is a larger scroll pump for those, but I can't justify the cost.
Consumer Products Guy wrote:
We had the very first 5971 system in our state. Our CEO strangely demanded that EVERY item ordered that year HAD TO arrive on site by Dec. 31, I actually asked Hewlett-Packard/Agilent to ship out an empty container, but they assured that it would arrive exactly on Dec. 31, and it did.
It ran on Windows 286 or 386 (can't remember) and I'd never seen a computer mouse before. That Windows version did not even have File Manager, has MS-DOS Executive.
The original vacuum pump for that unit was upgraded for free by the HP service engineer a few months later, and he gave me the old pump to take home to use on vehicle AC applications. Pump still works.
twranger wrote:
New update to old post- have to agree with lifetime of the 1.5s, we have been putting Edwards RV3 on the instruments, quieter and longer lifetime.
We just had a Agilent (Varian) DS-42 fail on a 5977, its 6 years old so no complaints. Looks like Agilent is now selling either a DS-40M or their new IDP-3 scroll pump. Anyone had experience with these models?
twranger wrote:twranger wrote:
New update to old post- have to agree with lifetime of the 1.5s, we have been putting Edwards RV3 on the instruments, quieter and longer lifetime.
We just had a Agilent (Varian) DS-42 fail on a 5977, its 6 years old so no complaints. Looks like Agilent is now selling either a DS-40M or their new IDP-3 scroll pump. Anyone had experience with these models?
Back to topic, I've also heard you can't put an RV3 on a diffusion pump mass spec. Any truth to that?
benhutcherson wrote:
Agreed with the general sentiment of using an uprated pump-as you said both for longevity, and in my experience better vacuum. At one time, when I was waiting on the RV3 to arrive for the Varian 300, I repurposed a MASSIVE 220V Edwards(either an older 8 or 12 series) that came with a now idle(and maybe never to run again) Finnigan LCQ Duo LC-MS. The set-up was a bit awkward and it was noisy, but that pump would pull it down like crazy and get me to 10^-8 repeatably.
LALman wrote:benhutcherson wrote:
Agreed with the general sentiment of using an uprated pump-as you said both for longevity, and in my experience better vacuum. At one time, when I was waiting on the RV3 to arrive for the Varian 300, I repurposed a MASSIVE 220V Edwards(either an older 8 or 12 series) that came with a now idle(and maybe never to run again) Finnigan LCQ Duo LC-MS. The set-up was a bit awkward and it was noisy, but that pump would pull it down like crazy and get me to 10^-8 repeatably.
My E2M2 gives me the following vacuums at the pump inlet on a 5971 (diffusion pump turned off) at various helium carrier flows (20m x 180um x 1mm DBVRX column.
Flow mL/min Vacuum mTorr
0.20 6.6
0.30 8.2
0.40 9.9
0.50 11.6
0.60 13.2
0.70 14.8
0.80 16.6
0.90 18.3
1.00 20.1
With the diffusion pump running and a carrier flow of 0.6mL/min I got 3.3 E-5 mTorr in the 5971.
Should I conclude my E2M2 is worn out? Should I get a RV3 do you think?