By Anonymous on Friday, June 20, 2003 - 01:53 am:

Which RP columns and mobile phase (pH) do you advise and why to separate these components creatine NH2-C(NH)-N(CH3)-CH2-COOH, guanidino acetate NH2-C(NH)-NH-CH2-COOH, L-Arginine NH2-C(NH)-NH-CH2-CH2-CH2-C(NH2)-COOH and creatinine CO-NH-C(NH)-N(CH3)-CH2.Thanks in advance.

-------------------------------------------------------------------------------------------------------
By Anonymous on Friday, June 20, 2003 - 05:53 pm:

I recommend a silica HILIC column with about 75% acetonitrile, 25% water with a control of the pH, maybe acetic acid around pH 4.75. The reason for this is that you will have a hard time to get retention for these compounds in reversed-phase without trying ion-pair, which creates difficulties with MS detection. HILIC is ideal for these very polar compounds. In addition, you get much higher sensitivity in ESI MS.

-------------------------------------------------------------------------------------------------------
By Martine Vaglio on Friday, September 5, 2003 - 06:30 am:

Thanks for the answer.
I receive an HILIC silica,3µm,50mm,id 2.1mm,column from Waters.
I use 0.1% formic acid (A) in water and acetonitrile (B) for the mobile phase with a flow rate of 0.2 ml/min.
I try an isocratic separation, the A/B 35:65 give me the best results. Also I try a gradient separation, the A/B 30:70 to A/B 45:55 in 15 minutes give me the best results, probably I could do it in a short time.
My questions are:
is acetic acid better than formic acid and why?
is 0.1% formic acid enough, is it better with 0.2%?
which percent is necessary to wash my column after the gradient separation A/B 50:50 or 40:60?
which separation is better to use isocratic or gradient and why.
Thanks to clear my troubles.

-------------------------------------------------------------------------------------------------------
By Anonymous on Saturday, September 6, 2003 - 06:01 pm:

Glad you are successful! Since you are successful, there is no need to change.
I had suggested ammonium acetate to prevent potentially excessive retention due to the amine functions on your analytes. I could have suggested ammonium formate at pH 3.75 as well. But if formic acid works, leave it at that.
Higher formic acid concentrations are better if you need to inject a lot of sample, but since you are working with MS, this is not likely. If you are happy with 0.1% stick with it. I would wash the column after the runs with 50:50. Finally, I do not see a reason to prefer isocratic over gradient, unless you need to work in high speed mode and want to avoid the column reequilibration time in the gradient.

-------------------------------------------------------------------------------------------------------
By Jakub Krijt on Tuesday, October 21, 2003 - 05:49 am:

hello,
i plan to set up a method for creatine and guanidinoacetate in plasma and tissues. Could you please write me what kind of MS/MS detector (ion trap, Q?) do you use and in which mode (MRM,SIM?).Do you perform a SPE or other sample preparation before injecting the samples on LC?
Thanks J.

-------------------------------------------------------------------------------------------------------
By Anonymous on Wednesday, January 28, 2004 - 05:23 am:

Between two samples are you doing a wash (which number of column volume) then an equilibration (which number of column volume)of the HILIC column after the gradient separation of the analytes or only an equilibration and finally a wash of the column at the end of the run.
Thanks for the answer.

-------------------------------------------------------------------------------------------------------
By Anonymous on Tuesday, July 6, 2004 - 07:15 am:

Hello,

Do you recommend any handbook on sample preparation for LC/MS bioanalysis. We work with plsama/blood/serum samples for pharmacokinetic studies.
Regards.

-------------------------------------------------------------------------------------------------------
By Uwe Neue on Thursday, July 22, 2004 - 04:18 pm:

Last anomymous: if you contact me, I'll send you an article published by us in the Handbook of Analytical Separations volume 4, Bioanalytical Separations.