Dear all,

Does anyone know if nitrogen can be used as carrier gas in replacement of Helium or Hydrogen?
In fact, I just arrived in a lab, and I have to work on a GC-MS (3800-Saturn 2200) that has been off for years. Since we can't get helium and are not able to work with hidrogen, I'm trying to use nitrogen as the carrier gas, but I get a very bad mass resolution, and I wonder if it's because of nitrogen use or a leak on GC...
Thank you for your answer!

Regards,
Charlotte
Generally speaking, you can use N2 as a carrier gas, but your resolution will suffer for doing so. No Hydrogen?!? What will you use for make-up gas?

Why the decision not to use hydrogen? It's actually a much better choice for a substitute carrier than nitrogen is for most applications.
Thanks,
DR
Image
You can do the chromatography with nitrogen - it will take longer and you will lose some sensitivity, but other wise it works fine. The problem is with the ion trap MS, which is set up only to work with helium at a specific residual pressure.

Peter
Peter Apps
If you use FID or TCD, nitrogen is fine. As already mentioned, resolution and sensitivity will suffer. With MS, you will likely KILL the masspec within hours.
You can get hydrogen cheap and pure. It's not that dangerous as often perceived. It (H2) works OK with mass filter (quadrupole) instruments. Ion traps, however, are designed to use He as damping medium.
If your need is rather urgent, we might be able to help.
Alexei Gapeev
Millis Scientific, Inc.
gapeev@millisscientific.com
Tel. 877-844-2635
Thank you for your answers.
I don't want to use Hydrogen because the GC-MS has been off for years and no one in the lab knows if the appartus is OK, or if it's been problems during the last uses.
Anyway, our gas supplier can finely find me a bottle of Helium so it's gonna be easier to start up the GC-MS, check it and work with it....
DR wrote:
Generally speaking, you can use N2 as a carrier gas, but your resolution will suffer for doing so. No Hydrogen?!? What will you use for make-up gas?

Why the decision not to use hydrogen? It's actually a much better choice for a substitute carrier than nitrogen is for most applications.


Dr:
Please, help me.
I will operate with a CP-3800 Varian GC / FID. Is it necessary to adjust it to replace the drag gas? I want to change from N2 to H2.
Search this forum or use your favorite search engine - there are many articles that talk about what to consider when making the switch from one carrier gas to another.

I am not an expert on this. In general, FID should be a good candidate for moving from N2 to H2 but you will have to increase flow rate and the result should be shorter run times, higher signal to noise ratio and better resolution.
Thanks,
DR
Image
7 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry