Mass axis off for ion 502 - Agilent 5975c

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I can't seem to get the mass axis correct for 502. Every tune I do ends up at 501.9. I tried swapping out the source but still get 501.9. Has anybody else seen this problem?
That is 502. If I recall correctly, the spec was +/- 0.1 but it has been a while. 502 is not that strong an ion anyway so you might be faced with too little signal as well but essentially you are there.

Best regards,

AICMM
The autotune functions will usually end once it is within +/- 0.1 of the target. If you really need it to be 502.0 you may have to tweak the Mass Gain and Mass Offset numbers manually. Another thing you may notice is that the mass axis will be slightly different between Scan and Profile in tune.

Put in into Manual tune and Profile mode and adjust the Mass Gain and Mass Offset numbers one click at a time both positive and negative until you get a feel as to how they affect the numbers then just tweak the settings until you get the desired results. Gain will affect the higher masses more than lower while Offset affects them all equally I believe(but that may be reversed).

AMUGain and AMUOffset are used to adjust your peak widths in the same manner.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry