I get confused while trying to understand the difference between Base peak chromatogram, TIC and selected ion current profile (mass chromatogram). And I cannot see what is, in practice, the difference in information provided by the different graph. Any help will be greatly appreciated.
-
- Posts: 232
- Joined: Tue Sep 07, 2004 7:03 pm
The BPC shows only the signal from the most intense mass in any given mass spectrum, plotted versus time. The TIC shows the sum of the signals from all masses in any given spectrum, plotted versus time. Selected ion current or extracted ion chromatogram (EIC or XIC) shows the signal from a single mass selected by the user, whether or not that mass is the most intense at any given time.
The BPC will often look similar to the TIC, but less noisy. However, chromatographic peaks consisting of signal at many masses can be understated by the BPC.
The BPC will often look similar to the TIC, but less noisy. However, chromatographic peaks consisting of signal at many masses can be understated by the BPC.
-
- Posts: 39
- Joined: Tue Feb 08, 2005 5:14 pm
Thank you to all. It is clear now.
4 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science