Dear all,
I am doing LC/MS analyses using capillary LC. I use 50 um ID Peeksil tubing. Most of the time this works fine. However, I sometimes get post-column obstructions. I find it difficult to find an explanation for these post-column obstructions. I can imagine getting pre-column obstruction, for instance due to sample impurities. Is there anyone that has an explanation for this phenomenon? My system setup is the following: Mobile phase A, 5 mM ammium acetate, mobile phase B, 97.5 % acetonitrile, 2.5 % water, 5 mM ammonium actetate. I run gradients from 10%B to 100% B. The flow rate is approximately 20 ul/min (I am using a flow splitter to get these low flow rates).

I hope someone has an explanation for this so I can change my setup and prevent these sometimes annoying obstructions.

Koen
Only speculation, but ammonium acetate is a pretty fair nutrient. Any chance you have some microbial contamination *inside* the column? I have no idea how you would confirm this, though. :(
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry