MSCHemist, it sounds like we have something similar to you in terms of this quick swap/backflush set up. Is yours an Agilent system?
A couple of years ago a service engineer from Agilent came by and told us that with experience they had found that having 4 psi He on the line isn’t the way to go. I think they had said 4 psi as a standard when setting up these quick swap fittings initially.
If this is relevant, you could contact Agilent or contact me directly and I will try to remember all the technicalities as to why. We made the changes they proposed and things more reliable

Kevin.
soccerdad007 wrote:
MSCHemist, it sounds like we have something similar to you in terms of this quick swap/backflush set up. Is yours an Agilent system?
A couple of years ago a service engineer from Agilent came by and told us that with experience they had found that having 4 psi He on the line isn’t the way to go. I think they had said 4 psi as a standard when setting up these quick swap fittings initially.
If this is relevant, you could contact Agilent or contact me directly and I will try to remember all the technicalities as to why. We made the changes they proposed and things more reliable

Kevin.

That is what is in the Agilent Quickswap manual and I have tested it and it works. I tried unhooking the column while having manual tune open looking at m/z 28, 32, and 18 and found no air coming in.

Also I keep the transferline at 280degC
at 4psi, 0.1mm ID 0.17m length that translates to 1.3 ml/min flow into the MSD. That is optimal when the column flow is 0.9-1.2ml/min
MSCHemist wrote:
soccerdad007 wrote:
MSCHemist, it sounds like we have something similar to you in terms of this quick swap/backflush set up. Is yours an Agilent system?
A couple of years ago a service engineer from Agilent came by and told us that with experience they had found that having 4 psi He on the line isn’t the way to go. I think they had said 4 psi as a standard when setting up these quick swap fittings initially.
If this is relevant, you could contact Agilent or contact me directly and I will try to remember all the technicalities as to why. We made the changes they proposed and things more reliable

Kevin.

That is what is in the Agilent Quickswap manual and I have tested it and it works. I tried unhooking the column while having manual tune open looking at m/z 28, 32, and 18 and found no air coming in.

Also I keep the transferline at 280degC
at 4psi, 0.1mm ID 0.17m length that translates to 1.3 ml/min flow into the MSD. That is optimal when the column flow is 0.9-1.2ml/min

Thanks!
18 posts Page 2 of 2

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry