Hi all,

Yesterday I connected a deactived fused silica in front of the DB-5ms column with "Unlimate union kits". The peaks broaden a little bit. Is it normal? Should I modifiy the original temp. progame? Or is it a gap inside the connector, leading the peaks broaden?

By the way, will the classic press-fit connector with resin sealing give a better result?

Thanks for all input!

With any kind of union, press fit, etc. YOu need to be sure that you have a good, clean, square cut on the column. Follow the instructions carefully or you will have dead volume and/or leaks.

My own esperience with unions is that I have always had some difficulty. Pressfits have worked well for me - even without resin (it depends on the columns I am using with the press fit -- some work better than others).

If you use resin - read and follow the instructions carefully.
I have been using precolumns with pressfit unions for a long time. In my experience they are tricky to assemble but if you do it right there is no need for resin (in fact I discourage its use since , for me at least, it has caused leaks on settling).
Now I use columns from Phenomenex that have a built-in precolumn.
Mike
3 posts Page 1 of 1

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry