I’m using a Thermo LTQ XL for this work, coupled to an Accela LC, but I also have a Thermo TSQ available which I’ve tried out with essentially the same result. Not doing chromatography at this stage; either injecting from autosampler without column or direct infusion from syringe pump.
A post-doc is building a molecule which I can’t describe in detail because of IP issues but broadly it is a Si-based cage anchor with 8 arms/appendages each of 10 residues terminating in an amide, each of mass about 1000; the overall mass of the molecule is about 10kDa. The synthesis results in mostly 8 arms being attached but apparently with some with a smaller number of arms attached with decreasing number as the number of arms reduces. We’ve had this confirmed by MALDI-TOF but we don’t have this onsite and samples have been going out to an external commercial lab, which clearly has cost implications, hence the desire to be able to pick this up in the LTQ.
The solubility is not great and my solution is about 1.2mg/ml ie roughly 0.1uM. At this concentration, it’s invisible in the LTQ; the mass limit of the MS is 4000m/z, so I’m relying on 3+ charged ions being present. I’ve tried a range of MS conditions and gone through the compound optimisation built into the LTQ software aiming at the 3x charged ion. I’ve tried in water and with 0.1% formic acid and even played with DMSO and m-NBA in the mix to manipulate the charging behaviour. Wondering if there was a problem with the MS, I’ve run lysozyme in the same conditions at the same mass concentration (lower molar concentration) and see a nice clear multiply-charged envelope, quite sufficient to calculate the molecular mass. I can also see quite clearly the appendage peptide or the cage molecule run individually at the same molar concentration.
Can anyone tell me a) if there’s some obvious reason I don’t know about that explains why this is invisible or b) if there is something I should be trying to make it ionise?
Hopefully getting a delivery of a Q-TOF in a few weeks, which will be able to tell me if the singly-charged ion is visible, but I’d rather see something before that plus it’s aggravating not understanding what’s going on.
Let me know if there’s anything else I can share (short of details of the molecule) which would be helpful.