im trying to prep a GC-MS for a upcoming installation of a pyrolyzer. However I cannot get the leak under control. The leak check of the system shows me a persistent 7-9 % leak on the system. The step I have taken so far:

- Check pressure of vacuum: Amazing, 37 mTorr foreline pressure and 5E-8 Torr high vac
- Check system with Restek leak detector: nothing
- Check gas lines with detector: nothing
- shut down MS, remove column, do bubble test: He is flowing
- plug up column end with septum, do leak test of GC: system is tight.
- Replace MS side ferrule: no improvement
- Check system with leak detector again: nothing
- refill cal standard: no improvement
- put system into manual tune, spray with "freeze 75" compartment by compartment, no correponding peak shows up

I am at the end of my wisdom here. I'm supposed to hand in my dissertation in spring and still have measurments to do. Someone please save me! :(
Found the problem finally. I was very close to opening the MS. Tried to increase split flow to see what happens with the leak. Turns out someone set the gas saver flow to 0 mL/min. So the system was pulling air through the split assembly. After I shut it on again the leak reduced to 0.3%.
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 0 registered, 1 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry