Hi All,

Naive question:

Would it be ok to use a column guard: 2.7 μm particle size, L × I.D. 5 mm × 2.1 mm with a column: 2.7 μm particle size, L × I.D. 5 cm × 3 mm

Thanks a lot,

K.
KevinBe wrote:
Hi All,

Naive question:

Would it be ok to use a column guard: 2.7 μm particle size, L × I.D. 5 mm × 2.1 mm with a column: 2.7 μm particle size, L × I.D. 5 cm × 3 mm

Thanks a lot,

K.


It will usually work ok to have small differences in the guard and analytical column on HPLC since most times they have a very short length of tubing between them. Even if they are the exact same sizes, there is still that small connecting tube of a different I.D. between them.
The past is there to guide us into the future, not to dwell in.
KevinBe wrote:
Hi All,

Naive question:

Would it be ok to use a column guard: 2.7 μm particle size, L × I.D. 5 mm × 2.1 mm with a column: 2.7 μm particle size, L × I.D. 5 cm × 3 mm

Thanks a lot,

K.

Don't bother about the different diameter. But 5 mm is rather long for a pre-column: The volume of your guard column is about 5 % of you analytical column, so you might expect small shift in retention time.
Try out and see (and tell us) what happens.
Jörg
Hi All,
So, I just got an Ascentis Express C18 from Sigma Aldrich. They literally say to run it at 1mL/min (which is very high for a 3mm diameter). They also say that you can run it without a guard because they are confident that you won't clog... Yet I decided to use a guard, just in case : and it works just fine!
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry