Hello everyone,

I am currently working on a method to determine PAH in urine with GCMS. My peaks are nice and sharp in the spiked samples. However, when I run standards, I get extensive tailing and -80% in peak area. Liner and septum are brand new. Can this be a matter of sample/standard solvent? Can anyone help?
If your spikes look good but your standards don't, check your solvent samples vs standards. Also, is this a splitless injection?
Samples and standards should be in matching solvents. If not the calibration in one solvent may not give accurate quantification for samples in another solvent.

Solvent affects the vaporization in the inlet, so that could be causing the tailing in the standards. For best results everything should be the same between standards and samples if possible.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry