VOC/BTEX method on GC/MS/HS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi everyone,
It's my fisrt post here but i come sometime to look for idea or some information.
But today I'm blocked and hope any kind of help.
We have bought a new GC/MS/HS system in replacement of an old VARIAN ion trap system.
The system is : GC 1300 - ISQ7000 - TRIPLUS500 from ThermoFisher.

We have started to transfer our method for VOC/BTEX analysis on the new device.
Thermo have provided the method they use in their laboratory.
We have try to use this one and adapt to our old method with a moderate success.
The VOC mix contain about 50 molecules. The method works pretty good for the first part of the chromatogram but in the second part who start with Toluene, xylenes elution time, problems appears :
Calibration becomes very quadratic and points are difficult to align on the calibration curve.
It's seems to be in relation with type of molecule, example no problem with bromoform or VOC but same situation with different degrees with benzene family.
Some information about the method :
It's ZB 624 PLUS column (20*0.18*1), same model we use with success on our old system.
Triplus500 is an headspace module with loop/valve. I've no experience with this system, the old system works with combipal and syringe.
I work in a SIM mode and use internal standard (in routine 1,2 dichloroethane D4 but some try with Toluene D8 or Chlorobenzene D5 have not correct the problem).
The calibration for each analytes is between 1ppb to 100 ppb.
If there is somebody with some kind of experience with this system and or analytes, I can detailed more the method.
At this moment i've try many changes but without success…

Charles
I have no experience with this headspace system but a couple of things come to mind.
Does Thermo specify the temperature for the loop/valve in addition to the headspace vial?
You say the problem compound responses are quadratic, are they decreasing with increased concentration?
What are your inlet conditions?
Hi, thanks for your anwser !

The temperature they use is 70°C for vial and 80° for valve/loop.
I've tried to increase (as in the old varian system) to 80°/85°C but with out success.
For the problematics compound the response increase with concentration.
I've select some molecules here to show the response :
Image
Image
Image
Image

In this system, column is directly connected to the HS module, injector is partially removed.
Only the gas part keep in place. So, no liner are use. The method works with contant flow of 1 ml/min - split 15 ml/min.


Charles
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry