I have a peak at m/z 40 and 45 when I make air spectrum and leak check fial in auto tune. How I can fixed this problem?
-
- Posts: 3475
- Joined: Mon Jan 07, 2013 8:54 pm
- Location: Western Kentucky
You can take a can of the compressed electronics duster which should give a mass in the 50-60m/z range and spray it around the seal to the vacuum chamber while scanning the instrument and watch for that ion to appear and you will pinpoint where the leak is. If you don't see a leak at the seal, then try where the column connects and where the vacuum lines connect. If you find the location of the leak then you can inspect to see if there is anything there that should not be. I have seen a large leak occur by simply finding a hair laying across the o-ring seal before.
The past is there to guide us into the future, not to dwell in.
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science