MS on SRI 8610C GC?

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello everyone, new here.

Has anyone added MS to an SRI 8610C GC?
There are three things the MSD needs from GC
1. The column routed through the hole on GC side. Carrier flow less than 1.5 ml/min
2. Transfer line heater and thermal sensor connected to the spare GC thermal zone.
3. Start line wire (open collector GC input) which should be grounded in moment the GC starts analysis.
2 posts Page 1 of 1