Hi guys,

Have anyone tried before flow injection analysis of LCMSMS to optimise the fragmentor voltage?

Is it that instead of a column, we connect a nut in between and run standard solutions at different fragmentor voltage?

Thank you for the replies.

Regards,
hts123
Yes. In flow injection analysis, you place an union instead of the column (or a tubing that directly connects the HPLC injector to the MS) and make injections of a certain volume of standard solution (typically the injection volume that you are going to use in LC-MS) at regular intervals while modifying parameters. This provides a gain of time compared to LC-M injections. This is different from MS infusion where you would infuse a solution at a low flow rate in the MS using a syringe pump.

Usually, people first start with MS infusion to optimize compound-dependent parameters (e.g. collision energy, fragmentor), then move to flow injection analysis to optimize source parameters (desolvation temperature and flow, ESI voltage etc.).
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry