ESI-MS source fragmentation/dimerisation

Posted: Thu Apr 08, 2021 7:33 pm
by yumifuku
Hello everyone

I'm analysing a synthesised sample using ESI-MS. A customer wants to know if it's starting material dimerised accidentally during the synthesis besides its product.
I've observed ions of starting material (M1) as [M1+H]+, [M1+Na]+ and sodium adduct of its dimer (M2) as [M2+Na]+. However, how can I distinguish [M2+Na]+ from [2M1+Na]+ due to an ion source dimerisation?
Other ions such as [M2+H]+ or [M2+K]+ are not observed and the intensity of [M2+Na]+ is relatively high. From total ion current seeing the sample might not be concentrated too much. The sample might be a mixture but at this moment I did not use any column.
I'm thankful if I can hear some ideas.

Re: ESI-MS source fragmentation/dimerisation

Posted: Tue Apr 13, 2021 9:08 am
by lmh
By MS alone it's extremely hard to be certain. A lot depends on how it is expected to dimerise chemically. Many chemical dimerisations will involve mass change, but not all. Non-covalent adduct dimers (2M+Na etc.) tend to be rather fragile, so if you can do gentle MS2 you may find they disappear. But it's very hard to distinguish a non-covalent adduct dimer from a very fragile covalent dimer. You may be able to reduce the amount of dimer by working at a lower concentration, or increasing source energies a bit, but you'll never be sure there is no dimer.
This is where chromatography can be very useful. Of course it's hard to prove a negative, but it's unlikely the dimer would have exactly the same retention time as the monomer, so if you find that the 2M+Na peak coelutes exactly with the M+H peak, it is probable that they are the same thing. Of course you'll need to convince yourself that the M+H peak isn't an in-source fragment from a pure dimer!
Really the easiest situation is where you do have both the dimer and the monomer and you can separate them with chromatography.

Re: ESI-MS source fragmentation/dimerisation

Posted: Tue Apr 13, 2021 7:15 pm
by yumifuku
Hi, Imh

Thank you for your reply. It sounds very interesting to measure gentle MSMS! I will try next time that. And I am also relieved to hear that it is difficult for others too.

Re: ESI-MS source fragmentation/dimerisation

Posted: Thu Apr 22, 2021 1:01 pm
by Ian_R
The proportion of [2M1+Na]+ will be higher the more concentrated the sample.
Assuming you are using flow injection (rather than constant infusion) see if the ratio of MNa+ to [2M1+Na]+ changes as the signal gets smaller. If it does then most likely what you are seeing is [2M1+Na]+
Agree that LC separation would be more definitive though.