There is something weird that happened to the GC-MS I am using. There is a small air leak since last week, and I working on it. However, I find something really wired, when I press the pre run and profile 28 N2 and 32 O2 and 18 H2O via manual tune, the amount of all those 3 ions crazily increased. and the total flow is less than column flow 1ml/min + purge flow 3min/l. The inlet is in splitless mode, 200C.
I really appreciate any suggestions.
Best
If the flows are reading incorrect numbers, maybe something is wrong the the flow controller and air is being pulled back into the inlet by the vacuum of the MS. If you remove the column and put in a ferrule with no hole in it and the air and water go away in the MS then it is probably something before the column.

If you don't have a ferrule with no hole, then you could remove the inlet end of the column and cap it by inserting it into a septa, the seal will be good but not perfect.
The past is there to guide us into the future, not to dwell in.
James_Ball wrote:
If the flows are reading incorrect numbers, maybe something is wrong the the flow controller and air is being pulled back into the inlet by the vacuum of the MS. If you remove the column and put in a ferrule with no hole in it and the air and water go away in the MS then it is probably something before the column.

If you don't have a ferrule with no hole, then you could remove the inlet end of the column and cap it by inserting it into a septa, the seal will be good but not perfect.


Thank you very much. After I isolated the MS I found the problem is from the GC part, so I took off the column, sealed the inlet and purge flow outlet, the pressure of the system dropped much faster than the manual marked. Then I realize either the EPC or the gas line has a problem.
I wonder how to check the flow valves of EPC and how to do the maintenance. Thank you again for your reply.
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry