very strange GCMS baseline

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

6 posts Page 1 of 1
Hi all,
I am using agilent 7890GC 5975MS for SPME analysis, scan range 28-200. recently I found small and sharp peaks in the chromatogram (sample and also blank) m/z is 28, they appear constantly at intervals through the whole chromatogram, I have no clear idea what's the problem.
Thank you in advance for your suggestions.
What does your tune look like? m/z=28 is nitrogen (N2). Perhaps you have a leak.

I recently put a helium tank on my systems and the tune gave a huge m/z=28. It was actually the base peak. I swapped out the tank and after an hour or 2 all went back to normal.
Unless I needed that low end of the scan range for some specific reason I'd raise the range up to avoid interference from N2 and O2
It is most likely trace nitrogen in the carrier. Since the lower limit of the scan range is set to 28, then if it reads the nitrogen peak at 27.5 it will reject it but at 28 and above it displays it. The best way to solve the problem is set the scan lower limit to 27 and you should have a smoother baseline just showing the trace nitrogen.

If you do not absolutely need to see 28m/z, then move the lower limit to about 35 for a much cleaner baseline.
The past is there to guide us into the future, not to dwell in.
James_Ball wrote:
It is most likely trace nitrogen in the carrier. Since the lower limit of the scan range is set to 28, then if it reads the nitrogen peak at 27.5 it will reject it but at 28 and above it displays it. The best way to solve the problem is set the scan lower limit to 27 and you should have a smoother baseline just showing the trace nitrogen.

If you do not absolutely need to see 28m/z, then move the lower limit to about 35 for a much cleaner baseline.

Thanks a lot, peaks disappeared after setting the limit to 27.
Thanks for all.
6 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry