In this article they used GC–MS and compared TIC peak areas to get relative concentration of molecules.

Why didn't they create EIC for the specific masses and measure those peak areas? If molecules are fragmented you may need to sum up areas of all the fragments (if the software can't do this for you). But the alternative (TIC) is that you may incorporate some co-eluting molecules which will lead to incorrect conclusions.

So when should TIC peaks used?
Software Engineer at elsci.io (my contact: stanislav.bashkyrtsev@elsci.io)
I have used it when analyzing for gasoline components. It would be difficult to sum all the ions in a hydrocarbon mixture. In that case the MS detector is used in the same manner as an FID.
Got it, thanks!

If others have other use cases, please chime in.
Software Engineer at elsci.io (my contact: stanislav.bashkyrtsev@elsci.io)
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry