Hi,

I have been given a sample of QAC (a mix of DDAC, ADBAC and ADEBAC) which was many orders of magnitude stronger than I was told. The upshot is that I have massively contaminated both the chromatogrphay side (Waters UPC2) and the fluidics on the MS (Xevo G2XS QTOF). I have been flushing for days with all manner of different solvents but the signals are still huge. Any suggestions?

cheers.
Have you tried a much higher concentration of Formic or Acetic acid in the solvents, either water or acetonitrile?
The past is there to guide us into the future, not to dwell in.
James_Ball wrote:
Have you tried a much higher concentration of Formic or Acetic acid in the solvents, either water or acetonitrile?



Hi,

Thank you for replying to my post. What sort of concentration are you thinking?

Thanks
3 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry