I have been approached by somebody in another lab with the request to identify peaks in their HPLC run by LC-MS. However, their mobile phase contains 30 mM sulfuric acid (pH 1.5)....
I am not keen on having that mobile phase go through our ESI source. and cause corrosion, especially at elevated temperature. At the same time, it is critical to identify said peaks.
Do you have any suggestions on how to compatibilize this HPLC method with ESI (or APCI)?
I am considering installing a 1:10 or so splitter post-column and then mixing with enough ammonium acetate buffer before the source to neutralize the acid. I will not be sensitivity limited at all. Does that sound like a viable plan?
Should I expect long term contamination with sulfuric acid-related ions? I had this happen with phosphoric acid before.
Thank you very much in advance!