I'm new with ICP-OES and I need help about the calibration curve.
I have an iCap 7200 ThermoFisher coupled with an autosampler ASX 560. I want to anticipate that I will buy ultrapure water and ultrapure acids; I avoid to use glass in my preparation, but I use only a glass flask to prepare the washing solution (I use a 20% acid solution composed by 10 ml HNO3 and 10 ml HCl in 1L total) because I haven't yet plastic flasks.
I prepared five standard solutions from a 100mg/L multielemental stock: 20 ppm, 10 ppm, 1 ppm, 0.1 ppm, 0.01 ppm. I explain my operations.
In the 50 ml graduated falcons I pipetted:
1) 20 ppm : 10 ml stock + 40 ml from washing solution (I helped with the graduated scores on the falcon)
2) 10 ppm: 5 ml stock + 45 ml washing solution
3) 1 ppm: 0.5 ml stock + 49.5 ml washing solution
4) 0.1 ppm: 0.060 ml (i.e. 60 um) stock in 60 ml total volume
5) 0.01 ppm: 5 ml from 4) + 45 ml washing solution.
My problem is that all concentrations are double after acquisition. Why?
I add that I had the same operations on the last week and the concentrations were good.
Can someone help me to understand my mistakes? Where is the problem?
Thank you!