I have inherited a 6890/5973 gcms that was used for VOC analyses and solvent extracts and analytical standards. I did not know the full history of the machine, but I did know that it had been awhile since it had been serviced. So I changed the column, liner, and septum.
So I went to run some hydrocarbon standards that were 100ng, but I am having problems. The peak of one compound got up to 18000 in abundance, but the baseline is very noisy and looks bad (similar to this: https://www.flickr.com/photos/129144858 ... 107490714/ )
I need my instrument to be able to detect <100ng (with a clean baseline ideally) and I'm not sure what to do.
Here are other details:
Column: db-5ms 30m, 0.25mm, 0.25um
inlet pressure:8.1
purge :10.0
toal:14.3
column flow:1.1
Autotune report from today:
Mass: 69 219 502
Ab: 386107 404042 55367
PW50: 0.60 0.61 0.61
Massgain:316 Repeller:27.78 MSsource:230
Massoffs:-11 Ionfcus:89.2 MSquad:150
Amugain:2223 Entlens:9.5 turbospeed:100
Amuoffs:122 Entoffs:19.33
Wid219:-0.048 EMVolts: 1576
h2o:0.35% n2:4.77% o2:1.43% co2:0.11%
Ramp criteria:
Ion Focus max: 90V using 502 ; EM gain 57358
Repeller max: 35V using ion 219; gain factor 0.47
Any advice would be very appreciated. Thank you.