Hi,
Accidentally we used a high concentration of ammonium acetate buffer for a LCMS with a triple quadrupole. I know this is not good for the column, but does it harm the quadrupole?
Very often I read the advice not to use more than 20mM. Where does this number come from? What could happen?
Thanks for help
Although it's volatile, it's not very very volatile, so if you pump too much in, it will start to block things, like the aperture at the front of the mass spec (ion transfer capillary or whatever it's called in your instrument). I don't know if it will also contaminate ion optics.
My inclination with your instrument in its current state would be to see if it still works, and if it does, breathe a sigh of relief and go back to normal concentrations. If there is a check-tune option, it might be worth doing it, as that's a quick way to check that it's a happy instrument. Alternatively just run a known standard and make sure the spectral quality is as good as normal, the peak the right size, etc.
good luck!
I've used concentrations of ammonium acetate up to 3x the 20mM recommendation to try and speed up methods on a QToF. I have not noticed any issues with sensitivity loss or the instrument going out of tune, however I make sure the source is clean before beginning a run.

As far as I know the 20 mM thing is more for ion formation using ESI, if you get above that then you start to lose ionization efficiency.
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry