GCMS- Tekmar 3000 dual concentrator. Naphthalene recoveries

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I have a dual concentrator system. The naphthalene recoveries are off. It reads high on concentrator 1, but low on concentrator 2. The 6- port valves and sparge mounts have been cleaned, I replaced the sparge vessels, and the tubing has been replaced. I also changed the traps on both concentrators. A new calibration curve was also run. What else could be the problem?

Thanks for any input!
Do you know if the flow and temperatures of both concentrators are the same?

Possibly try increasing the transfer line temperature of the Low concentrator 5-10C and see if that makes a difference. Also if you have heater jackets for the sparge tubes you could set them to 40C and make sure there are no temperature differences between the sparge tubes due to drafts. A HVAC duct hitting one sparge tube and cooling it could reduce the purgeability of Naphthalene.
The past is there to guide us into the future, not to dwell in.
2 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry